Journal ArticleDOI
TL;DR: An ultrahigh-speed spectral domain optical coherence tomography (SD-OCT) system is presented, demonstrating the ability to acquire high-quality structural images with an axial resolution of 6 microm at ultrahigh speed and with an ocular exposure level of less than 600 microW.
Abstract: An ultrahigh-speed spectral domain optical coherence tomography (SD-OCT) system is presented that achieves acquisition rates of 29,300 depth profiles/s. The sensitivity of SD-OCT and time domain OCT (TD-OCT) are experimentally compared, demonstrating a 21.7-dB improvement of SD-OCT over TD-OCT. In vivo images of the human retina are presented, demonstrating the ability to acquire high-quality structural images with an axial resolution of 6 microm at ultrahigh speed and with an ocular exposure level of less than 600 microW.

643 citations


Journal ArticleDOI
TL;DR: A white-light interference microscope for ultrahigh-resolution full-field optical coherence tomography of biological media and subcellular-level images of plant, animal, and human tissues are presented.
Abstract: We have developed a white-light interference microscope for ultrahigh-resolution full-field optical coherence tomography of biological media. The experimental setup is based on a Linnik-type interferometer illuminated by a tungsten halogen lamp. En face tomographic images are calculated by a combination of interferometric images recorded by a high-speed CCD camera. Spatial resolution of 1.8 μm × 0.9 μm (transverse × axial) is achieved owing to the extremely short coherence length of the source, the compensation of dispersion mismatch in the interferometer arms, and the use of relatively high-numerical-aperture microscope objectives. A shot-noise-limited detection sensitivity of 90 dB is obtained in an acquisition time per image of 4 s. Subcellular-level images of plant, animal, and human tissues are presented.

483 citations


Journal ArticleDOI
TL;DR: In this article, a closed-loop correction of both ocular and system aberrations results in a residual uncorrected wavefront rms of 0.1 µm for a 3.68 mm pupil diameter.
Abstract: Merging of ultrahigh-resolution optical coherence tomography (UHR OCT) and adaptive optics (AO), resulting in high axial (3 µm) and improved transverse resolution (5–10 µm) is demonstrated for the first time to our knowledge in in vivo retinal imaging. A compact (300 mm×300 mm) closed-loop AO system, based on a real-time Hartmann–Shack wave-front sensor operating at 30 Hz and a 37-actuator membrane deformable mirror, is interfaced to an UHR OCT system, based on a commercial OCT instrument, employing a compact Ti:sapphire laser with 130-nm bandwidth. Closed-loop correction of both ocular and system aberrations results in a residual uncorrected wave-front rms of 0.1 µm for a 3.68-mm pupil diameter. When this level of correction is achieved, OCT images are obtained under a static mirror configuration. By use of AO, an improvement of the transverse resolution of two to three times, compared with UHR OCT systems used so far, is obtained. A significant signal-to-noise ratio improvement of up to 9 dB in corrected compared with uncorrected OCT tomograms is also achieved.

345 citations


Journal ArticleDOI
TL;DR: A spatially adaptive two-dimensional wavelet filter is used to reduce speckle noise in time-domain and Fourier-domain optical coherence tomography (OCT) images.
Abstract: A spatially adaptive two-dimensional wavelet filter is used to reduce speckle noise in time-domain and Fourier-domain optical coherence tomography (OCT) images. Edges can be separated from discontinuities that are due to noise, and noise power can be attenuated in the wavelet domain without significantly compromising image sharpness. A single parameter controls the degree of noise reduction. When this filter is applied to ophthalmic OCT images, signal-to-noise ratio improvements of >7 dB are attained, with a sharpness reduction of <3%.

289 citations


Journal ArticleDOI
TL;DR: An analysis for polarization-sensitive optical coherence tomography that facilitates the unrestricted use of fiber and fiber-optic components throughout an interferometer and yields sample birefringence, diattenuation, and relative optic axis orientation is presented.
Abstract: We present an analysis for polarization-sensitive optical coherence tomography that facilitates the unrestricted use of fiber and fiber-optic components throughout an interferometer and yields sample birefringence, diattenuation, and relative optic axis orientation. We use a novel Jones matrix approach that compares the polarization states of light reflected from the sample surface with those reflected from within a biological sample for pairs of depth scans. The incident polarization alternated between two states that are perpendicular in a Poincare sphere representation to ensure proper detection of tissue birefringence regardless of optical fiber contributions. The method was validated by comparing the calculated diattenuation of a polarizing sheet, chicken tendon, and muscle with that obtained by independent measurement. The relative importance of diattenuation versus birefringence to angular displacement of Stokes vectors on a Poincare sphere was quantified.

279 citations


Journal ArticleDOI
TL;DR: The possibility of measuring a full Doppler flow depth profile in parallel by use of frequency-domain optical coherence tomography is demonstrated and the method is based on a local phase analysis of the backscattered signal and allows for imaging of bidirectional Dopplers flow.
Abstract: The possibility of measuring a full Doppler flow depth profile in parallel by use of frequency-domain optical coherence tomography is demonstrated. The method is based on a local phase analysis of the backscattered signal and allows for imaging of bidirectional Doppler flow. The Doppler frequency limit is 5 kHz for the presented measurements and is set by half of the frame rate of the CCD detector array. We measured the flow of 0.3‐µm microspheres suspended in distilled water at controlled flow rates and in vitro human blood flow through a 200‐µm capillary with a real-time color-encoded Doppler tomogram rate of 2–3/s.

264 citations


Journal ArticleDOI
TL;DR: Two types of ultrasound-mediated biophotonic imaging are reviewed–acousto-optical tomography and photo-acoustic tomography–both of which are based on non-ionizing optical and ultrasonic waves.
Abstract: This article reviews two types of ultrasound-mediated biophotonic imaging–acousto-optical tomography (AOT, also called ultrasound-modulated optical tomography) and photo-acoustic tomography (PAT, also called opto-acoustic or thermo-acoustic tomography)–both of which are based on non-ionizing optical and ultrasonic waves. The goal of these technologies is to combine the contrast advantage of the optical properties and the resolution advantage of ultrasound. In these two technologies, the imaging contrast is based primarily on the optical properties of biological tissues, and the imaging resolution is based primarily on the ultrasonic waves that either are provided externally or produced internally, within the biological tissues. In fact, ultrasonic mediation overcomes both the resolution disadvantage of pure optical imaging in thick tissues and the contrast and speckle disadvantages of pure ultrasonic imaging. In our discussion of AOT, the relationship between modulation depth and acoustic amplitude is clarified. Potential clinical applications of ultrasound-mediated biophotonic imaging include early cancer detection, functional imaging, and molecular imaging.

249 citations


Journal ArticleDOI
TL;DR: In this article, a high-speed dynamic focus control system based on a microelectromechanical mirror was designed to maintain high transverse resolution over the entire depth scan, which can yield sub-optimal lateral resolution outside the focal zone for deep-imaging OCT systems.

239 citations


Journal ArticleDOI
TL;DR: A miniature endoscope that is capable of rapid lateral scanning and is suitable for real-time forward-imaging optical coherence tomography (OCT) and operated in the new image acquisition sequence at 6 frames/s is demonstrated.
Abstract: We developed a miniature endoscope that is capable of rapid lateral scanning and is suitable for real-time forward-imaging optical coherence tomography (OCT). The endoscope has an outer diameter of 2.4 mm, consisting of a miniature tubular lead zirconate titanate (PZT) actuator, a single-mode fiber-optic cantilever, and a graded-index lens. Rapid lateral scanning at 2.8 kHz is achieved when the fiber-optic cantilever is resonated with the PZT actuator. This allows OCT imaging to be performed by fast lateral beam scanning followed by slow depth scanning, which is different from the conventional OCT imaging sequence. Real-time OCT imaging with the endoscope operated in the new image acquisition sequence at 6 frames/s is demonstrated.

230 citations


Journal ArticleDOI
TL;DR: A distally actuated, rotational-scanning micromotor endoscope catheter probe is demonstrated for ultrahigh-resolution in vivo endoscopic optical coherence tomography (OCT) imaging and promises to improve OCT imaging performance in future endoscopic imaging applications.
Abstract: A distally actuated, rotational-scanning micromotor endoscope catheter probe is demonstrated for ultrahigh-resolution in vivo endoscopic optical coherence tomography (OCT) imaging. The probe permits focus adjustment for visualization of tissue morphology at varying depths with improved transverse resolution compared with standard OCT imaging probes. The distal actuation avoids nonuniform scanning motion artifacts that are present with other probe designs and can permit a wider range of imaging speeds. Ultrahigh-resolution endoscopic imaging is demonstrated in a rabbit with <4-microm axial resolution by use of a femtosecond Cr:forsterite laser light source. The micromotor endoscope catheter probe promises to improve OCT imaging performance in future endoscopic imaging applications.

210 citations


Journal ArticleDOI
TL;DR: A novel endoscopic optical coherence tomography probe was designed and constructed with a 1.9-mm microelectromechanical system (MEMS) motor that does not require the coupling of a rotational single-mode fiber and has the advantages of having a much smaller diameter and being more flexible than traditional endoscopes.
Abstract: A novel endoscopic optical coherence tomography probe was designed and constructed with a 1.9-mm microelectromechanical system (MEMS) motor. The new MEMS endoscopic probe design eliminates the need to couple the rotational energy from the proximal to the distal end of the probe. Furthermore, the endoscopic probe's sheath and fiber have the advantages of having a much smaller diameter and being more flexible than traditional endoscopes since no reinforcement is needed to couple the rotational torque. At the distal end, a prism mounted on a micromotor deflects the light rays to create a transverse circular-scanning pathway. Because our MEMS scanner does not require the coupling of a rotational single-mode fiber, a high scanning speed is possible while eliminating unstable optical signals caused by nonuniform coupling.

Journal ArticleDOI
TL;DR: The experimental implementation of this novel concept of noncontact optical measurements from diffuse media is described and capacity to perform small animal imaging is demonstrated.
Abstract: Noncontact optical measurements from diffuse media could facilitate the use of large detector arrays at multiple angles that are well suited for diffuse optical tomography applications. Such imaging strategy could eliminate the need for individual fibers in contact with tissue, restricted geometries, and matching fluids. Thus, it could significantly improve experimental procedures and enhance our ability to visualize functional and molecular processes in vivo. In this paper, we describe the experimental implementation of this novel concept and demonstrate capacity to perform small animal imaging.

Journal ArticleDOI
TL;DR: Experimental and simulation studies led to the conclusion that objects can be automatically detected with CNR analysis and that the current system has a spatial resolution limit near 4 mm and a contrast resolution limit close to 1.4.
Abstract: The contrast-to-noise ratio (CNR) was used to determine the detectability of objects within reconstructed images from diffuse near-infrared tomography. It was concluded that there was a maximal value of CNR near the location of an object within the image and that the size of the true region could be estimated from the CNR. Experimental and simulation studies led to the conclusion that objects can be automatically detected with CNR analysis and that our current system has a spatial resolution limit near 4 mm and a contrast resolution limit near 1.4. A new linear convolution method of CNR calculation was developed for automated region of interest (ROI) detection.

Journal ArticleDOI
TL;DR: It is demonstrated theoretically and with a human experiment that a simple geometry of sources and detectors can provide overlapping measurements within the limitation of instrumentation dynamic range and produce an image resolution and localization accuracy that is twofold better.
Abstract: We compare two geometries of sources and detectors for optimizing the diffuse optical imaging resolution of brain activation in humans. Because of limitations in the instruments' dynamic range, most diffuse optical brain activation images have used only nonoverlapping measurements. We demonstrate theoretically and with a human experiment that a simple geometry of sources and detectors can provide overlapping measurements within the limitation of instrumentation dynamic range and produce an image resolution and localization accuracy that is twofold better.

Journal ArticleDOI
TL;DR: The results show that adaptive techniques enable the computationally efficient and stable solution of the inverse imaging problem while providing the resolution necessary for imaging the signals from molecularly targeting agents.
Abstract: A three-dimensional fluorescence-enhanced optical tomography scheme based upon an adaptive finite element formulation is developed and employed to reconstruct fluorescent targets in turbid media from frequency-domain measurements made in reflectance geometry using area excitation illumination. The algorithm is derived within a Lagrangian framework by treating the photon diffusion model as a constraint to the optimization problem. Adaptively refined meshes are used to separately discretize maps of the forward/adjoint variables and the unknown parameter of fluorescent yield. A truncated Gauss-Newton method with simple bounds is used as the optimization method. Fluorescence yield reconstructions from simulated measurement data with added Gaussian noise are demonstrated for one and two fluorescent targets embedded within a 512ml cubical tissue phantom. We determine the achievable resolution for the area-illumination/area-detection reflectance measurement geometry by reconstructing two 0.4cm diameter spherical targets placed at at a series of decreasing lateral spacings. The results show that adaptive techniques enable the computationally efficient and stable solution of the inverse imaging problem while providing the resolution necessary for imaging the signals from molecularly targeting agents.

Journal ArticleDOI
TL;DR: Local polarization properties are mapped for the first time to the knowledge by using polar decomposition in combination with least-squares fitting to differentiate measured integrated Jones matrices with respect to depth.
Abstract: A unique feature of polarization-sensitive Mueller optical coherence tomography is that, by measuring Jones or Mueller matrices, it can reveal the complete polarization properties of biological samples, even in the presence of diattenuation. We map local polarization properties for the first time to our knowledge by using polar decomposition in combination with least-squares fitting to differentiate measured integrated Jones matrices with respect to depth. We also introduce the new concept of dual attenuation coefficients to characterize diattenuation per unit infinitesimal length in tissues. We experimentally verify the algorithm using measurements of a section of porcine tendon and the septum of a rat heart.

Journal ArticleDOI
TL;DR: In this paper, a polarization-sensitive Fourier domain optical coherence tomography (PS-FD-OCT) system is demonstrated, which is based on a spectral interferometer and enables phase information to be used in an OCT image.
Abstract: A polarization-sensitive Fourier domain optical coherence tomography (PS-FD-OCT) system is demonstrated. This OCT system is based on a spectral interferometer, does not require mechanical axial scanning, and enables phase information to be used in an OCT image. Owing to this phase information, the system requires only two measurements for determining the Jones matrix images and Muller images of biological samples. This system reveals the birefringence properties of the inner surface of a porcine esophagus.

Journal ArticleDOI
TL;DR: Induced haemodynamic and blood oxygenation changes occurring within the brain of a ventilated newborn infant have been imaged in three dimensions using optical tomography, showing large changes in absorption coefficient at both wavelengths.
Abstract: Induced haemodynamic and blood oxygenation changes occurring within the brain of a ventilated newborn infant have been imaged in three dimensions using optical tomography. Noninvasive measurements of the flight times of transmitted light were acquired during illumination of the brain by laser pulses at wavelengths of 780 nm and 815 nm. The oxygen and carbon dioxide partial pressures were adjusted through alterations to the ventilator settings, resulting in changes to the cerebral blood volume and oxygenation. Three-dimensional images were generated using the physiologically associated differences in the measured data, obviating the need for data calibration using a separate reference measurement. The results exhibit large changes in absorption coefficient at both wavelengths. Images corresponding to differences in concentrations of oxy- and deoxyhaemoglobin are in qualitative agreement with known physiological data.

Journal ArticleDOI
TL;DR: Near-infrared (NIR) dyes with absorption spectra features within the OCT source spectrum can be used for enhancing contrast in this situation and a link between SOCT and fluorescence imaging is demonstrated.
Abstract: Optical coherence tomography (OCT) images of biological tissues often have low contrast. Spectroscopic optical coherence tomography (SOCT) methods have been developed to enhance contrast but remain limited because most tissues are not spectrally active in the frequency bands of laser sources commonly used in OCT. Near-infrared (NIR) dyes with absorption spectra features within the OCT source spectrum can be used for enhancing contrast in this situation. We introduce and demonstrate the use of NIR dyes as contrast agents for SOCT. Contrast-enhanced images are compared with fluorescence microscopy, demonstrating a link between SOCT and fluorescence imaging.

Journal ArticleDOI
TL;DR: A novel optical tomographic imaging system that was designed to determine two-dimensional spatial distribution of optical properties in a sagittal plane through finger joints is presented and used to obtain tomographic images of six proximal interphalangeal finger joints from two patients with rheumatoid arthritis.
Abstract: We present a novel optical tomographic imaging system that was designed to determine two-dimensional spatial distribution of optical properties in a sagittal plane through finger joints. The system incorporates a single laser diode and a single silicon photodetector into a scanning device that records spatially resolved light intensities as they are transmitted through a finger. These data are input to a model-based iterative image reconstruction (MOBIIR) scheme, which uses the equation of radiative transfer (ERT) as a forward model for light propagation through tissue. We have used this system to obtain tomographic images of six proximal interphalangeal finger joints from two patients with rheumatoid arthritis. The optical images were compared to clinical symptoms and ultrasound images.

Journal ArticleDOI
TL;DR: System design and image reconstruction techniques are described, along with simulation and phantom results that demonstrate the characteristics and limitations of system accuracy and resolution.
Abstract: Laminar optical tomography (LOT) is a new technique that combines the advantages of diffuse optical tomography image reconstruction and a microscopy-based setup to allow noncontact imaging with 100–200‐µm resolution effective over depths of 0–2.5 mm. LOT is being developed primarily for multispectral imaging of rat cortex, for which resolving functional dynamics in various layers of the brain’s cortex (to depths of 1500 µm) is of increasing interest to neurophysiologists. System design and image reconstruction techniques are described, along with simulation and phantom results that demonstrate the characteristics and limitations of system accuracy and resolution.

Journal ArticleDOI
TL;DR: Experimental results with beef tendon and rabbit tendon and muscle show that this technique has promising potential for imaging the birefringent structure of multiple-layer samples with varying optical axes.
Abstract: Conventional polarization-sensitive optical coherence tomography (PS-OCT) can provide depth-resolved Stokes parameter measurements of light reflected from turbid media. A new algorithm that takes into account changes in the optical axis is introduced to provide depth-resolved birefringence and differential optical axis orientation images by use of fiber-based PS-OCT. Quaternion, a convenient mathematical tool, is used to represent an optical element and simplify the algorithm. Experimental results with beef tendon and rabbit tendon and muscle show that this technique has promising potential for imaging the birefringent structure of multiple-layer samples with varying optical axes.

Journal ArticleDOI
TL;DR: A technique for detection of ultrasound-induced optical phase modulation using an adaptive, photorefractive-crystal-based interferometry system and experimental results are presented demonstrating detection of ultrasound-modulated signals in highly scattering media by use of pulsed ultrasound insonation.
Abstract: Ultrasound-modulated optical tomography is a dual-wave sensing technique in which diffusive light in a turbid medium interacts with an imposed acoustic field. A phase-modulated photon field emanates from the interaction region and carries with it information about the optomechanical properties of the medium. We present a technique for detection of ultrasound-induced optical phase modulation using an adaptive, photorefractive-crystal-based interferometry system. Experimental results are presented demonstrating detection of ultrasound-modulated signals in highly scattering media by use of pulsed ultrasound insonation.

Journal ArticleDOI
TL;DR: A multimodality system that combines optical coherence tomography (OCT) and laser-induced fluorescence (LIF) in a 2.0-mm-diameter endoscopic package is designed and the use of a rod prism is presented to reduce the asymmetry in the OCT beam caused by a cylindrical window.
Abstract: We have designed a multimodality system that combines optical coherence tomography (OCT) and laser-induced fluorescence (LIF) in a 2.0-mm-diameter endoscopic package. OCT provides ∼18-μm resolution cross-sectional structural information over a 6-mm field. LIF spectra are collected sequentially at submillimeter resolution across the same field and provide histochemical information about the tissue. We present the use of a rod prism to reduce the asymmetry in the OCT beam caused by a cylindrical window. The endoscope has been applied to investigate mouse colon cancer in vivo.

Journal ArticleDOI
TL;DR: In this paper, a 1-mm/sup 2/2/2 2/3/2-axis, single-crystalline-silicon (SCS)-based aluminum-coated scanning micromirror with large rotation angle (up to 40/spl deg/), which can be used in an endoscopic optical coherence tomography imaging system is presented.
Abstract: This paper reports a 1-mm/sup 2/, two-axis, single-crystalline-silicon (SCS)-based aluminum-coated scanning micromirror with large rotation angle (up to 40/spl deg/), which can be used in an endoscopic optical coherence tomography imaging system. The micromirror is fabricated using a deep reactive ion etch post-CMOS micromachining process. The static response, frequency response, resonance frequency shift, and thermal imaging of the device are presented. A 4/spl times/4 pixel display using this two-dimensional micromirror device has been demonstrated.

Journal ArticleDOI
TL;DR: This procedure imposes prior spectral information into the image reconstruction that results in a dramatic improvement in the image contrast-to-noise ratio of better than 100% and is demonstrated with simulations and a dynamic blood phantom experiment.
Abstract: We present an algorithm to reconstruct chromosphere concentration images directly rather than following the traditional two-step process of reconstructing wavelength-dependent absorption coefficient images and then calculating chromosphere concentration images. This procedure imposes prior spectral information into the image reconstruction that results in a dramatic improvement in the image contrast-to-noise ratio of better than 100%. We demonstrate this improvement with simulations and a dynamic blood phantom experiment.

Journal ArticleDOI
TL;DR: As the resolution of clinical OCT systems improves, the capability to reproducibly map complex structures in the living eye at high resolution will lead to improved understanding of disease processes and improved sensitivity and specificity of diagnostic procedures.
Abstract: An experimental tracking optical coherence tomography (OCT) system has been clinically tested. The prototype instrument uses a secondary sensing beam and steering mirrors to compensate for eye motion with a closed-loop bandwidth of 1 kHz and tracking accuracy, to within less than the OCT beam diameter. The retinal tracker improved image registration accuracy to <1 transverse pixel (<60 microm). Composite OCT images averaged over multiple scans and visits show a sharp fine structure limited only by transverse pixel size. As the resolution of clinical OCT systems improves, the capability to reproducibly map complex structures in the living eye at high resolution will lead to improved understanding of disease processes and improved sensitivity and specificity of diagnostic procedures.

Journal ArticleDOI
TL;DR: In this article, the authors describe the methods used and focus on their implementation in such a universal software, which focuses on high-resolution digital fringe evaluation including phase stepping, Fourier domain evaluation as well as unwrapping techniques for regular and irregular fringe patterns.

Journal ArticleDOI
TL;DR: In this paper, a multispectral, frequency-domain near infrared tomography (NIR) system was proposed for breast tissue optical absorption and reduced scattering coefficients using a Newton-type reconstruction algorithm.
Abstract: The design and implementation of a multispectral, frequency-domain near infrared tomography system is outlined, which operates in a MRI magnet for utilization of MR-guided image reconstruction of tissue optical properties. Using long silica optical fiber bundles, measurements of light transmission through up to 12 cm of female breast tissue can be acquired simultaneously with MRI scans. The NIR system utilizes six optical wavelengths from 660 to 850 nm using intensity modulated diode lasers nominally working at 100 MHz. Photomultiplier tube detector gain levels are electronically controlled on a time scale of 200 ms, thereby allowing rapid switching of the source to locations around the tissue. There are no moving parts in the detection channels and for each source position, 15 PMTs operating in parallel allow sensitivity down to 0.5 pW/cm2 at the tissue surface. Images of breast tissue optical absorption and reduced scattering coefficients are obtained using a Newton-type reconstruction algorithm to solv...

Journal ArticleDOI
TL;DR: A novel implementation of high-resolution ultrasound-modulated optical tomography that, based on optical contrast, can image several millimeters deep into soft biological tissues, and has the potential for broad biomedical applications.
Abstract: We present a novel implementation of high-resolution ultrasound-modulated optical tomography that, based on optical contrast, can image several millimeters deep into soft biological tissues. A long-cavity confocal Fabry–Perot interferometer, which provides a large etendue and a short response time, was used to detect the ultrasound-modulated coherent light that traversed the scattering biological tissue. Using 15-MHz ultrasound, we imaged with high-contrast light-absorbing structures placed > 3 mm below the surface of chicken breast tissue. The resolution along the axial and the lateral directions with respect to the ultrasound propagation direction was better than 70 and 120 µm, respectively. The resolution can be scaled down further by use of higher ultrasound frequencies. This technology is complementary to other imaging technologies, such as confocal microscopy and optical-coherence tomography, and has the potential for broad biomedical applications.